The Lab Savior: A Comprehensive HPLC Troubleshooting Guide

09 March 2026

Quick-Start Diagnostic Summary 

If you are currently facing an instrument issue, follow this logical flow to isolate the source: Mobile Phase → Pump → Injector/Autosampler → Column → Detector.  

  • No Signal? Verify the detector lamp is ON, check the wavelength setting, and purge the pump to remove air. 
  • Pressure Anomalies? High pressure usually indicates a blocked frit or column; low pressure often suggests a leak or air in the pump heads.
  • Peak Shape Issues? Address split peaks by replacing the guard column; resolve fronting by reducing injection volume; check fittings for dead volume if peaks are tailing. 
  • Retention Time (RT) Shifts? Stabilize the column oven temperature and ensure the mobile phase is fresh and degassed.
  • Expert Maintenance: Contact SeqGen for persistent Sciex MS sensitivity loss, internal Shimadzu failures, or annual Preventive Maintenance (PM)/compliance requirements. 

Read More

Troubleshooting Vacuum Pressure Issues on the SCIEX Triple Quad 6500+

20 June 2025

Not reaching “Ready” on your SCIEX Triple Quad 6500+? Vacuum pressure could be the culprit. Seeing flashing LEDs, high Torr readings, or sudden loss of sensitivity? You’re not alone—and the fix may be simpler than you think. This guide answers common questions about vacuum issues on the SCIEX 6500+ and walks you through quick checks to get your system back online.

Read More

Perform routine maintenance for mobile phases in the LCMS system

11 April 2023

When it comes to liquid chromatography, the mobile phase plays a critical role in carrying the sample through the chromatography column and into the mass spectrometer.

Read More

How to save Windows event logs in SCIEX Analyst software?

09 March 2023

SCIEX Triple Quad and QTRAP systems are well known for their sensitivity and robustness in mass detection and quantification. But even with modern design, things could go wrong out of nowhere.

Read More

Troubleshooting Mass Spectrometers 4500/5500/6500

17 February 2022

Having issues activating your Mass Spectrometer profile?

Our expert team has helped many users to activate the profile in the hardware configuration. Troubleshooting your Mass Spectrometer is essential to figure out any issue when you cannot activate it. 

Read More

Subscribe to Email Updates

New Call-to-action 

Recent Posts

SeqGen is an independently owned and operated provider of repair services and refurbished laboratory equipment. Any product names, logos, brands, and other trademarks or images featured or referred to within the SeqGen website and/or on any social media forum are the property of their respective trademark holders or OEMs. These trademark holders are not affiliated with SeqGen or its website. These trademark holders do not sponsor or endorse SeqGen or any of its products or comments. The use of any trade name or trademark is for identification and reference purposes only and does not imply any association with the trademark holder of their product brand.